limited memory

The dataTaker DT85 16 channel general purpose data logger

limited memory

Postby jodieugent » Fri Feb 16, 2018 4:10 am

For are thesis we are doing research to the hygrothermal behavior of wooden walls. To make measurements we are using the datataker dt85. Our problem is now that the logger only gives the results for approximately 10 days, what's about 10 000 data, while in the manual of the datalogger is written that it has a memory for about 10 million data. Can we change anything to the setting so we can collect data for a longer period in one time?

Thank you in advance.
jodieugent
Newbie
 
Posts: 1
Joined: Fri Feb 16, 2018 4:00 am

Re: limited memory

Postby Roger » Mon Feb 19, 2018 7:51 pm

Good evening jordeugent
You can control the amount of data for each schedule.
Where you define the schedule click on the storage tab and set the data stroage to the size you require

Cheers
Roger
"Don't Panic"
Douglas Adams HHGTTG
Roger
Guru
 
Posts: 286
Joined: Fri Feb 10, 2012 10:38 am